TDS3014B connecting with and RS232 to USB adapter

Moderators: Hiker, notnomis, Buck, Super Mod

TDS3014B connecting with and RS232 to USB adapter

Postby sscribbins on Mon Jun 27, 2011 12:12 pm

Hi all

I have a TDS 3024B which I am connecting to a computer which has only a single RS232 port on it. I now have another device which requires the serial port on the PC, so I need to connect the scope via a serial to USB adapter. I know this is possible, as I have seen it done before with the exact same device but I seem to be having some trouble configuring the software to communicate with the scope.

When connecting to the scope via serial there is no problem, but as soon as I switch to the usb adapter, the Instrument Manager Service software can see the scope but a connection test results in the error message:
"Device Proxy (tek TDS3014) is running but the attached hardware is not responding. Response string is: ()." (SW Status: Running, HW Status: Not Responding on the instrument manager main screen).

I notice that the scope is a 3024 and the software sees a 3014, but this works fine in serial (I don't know how much difference there is between the 3014 and 3024, if any, so I have no idea if this is important - thought I'd mention it just in case it's relevant).

Can anyone help with this?

Thanks

S
sscribbins
 
Posts: 2
Joined: Mon Jun 27, 2011 11:47 am

Re: TDS3014B connecting with and RS232 to USB adapter

Postby cball on Mon Jun 27, 2011 11:00 pm

The tds3012B and TDS3022B is only different in bandwidth. The connection however should be the same. USB-RS232 adapters are going to assign the new resource to a different COM port that the physical port on your computer. Make sure you have the correct one selected. Also, is the adapter Null Modem?
Tektronix Application Engineer
Anything I post here is user to user advice. For official technical support:
http://www.tek.com/webform/technical-support-request
cball
 
Posts: 575
Joined: Fri Dec 17, 2010 11:17 pm
Location: Beaverton, OR

Re: TDS3014B connecting with and RS232 to USB adapter

Postby BOBAH on Sun Dec 18, 2011 4:01 am

I have exactly the same error message !!
Please help.

I installed a PCI-GPIB car on a new machine with Windows 7 (32 bit)
I installed NI-488.2 from NI site.
When I run Measurement and Automation (from NI) I can see my scope TDS 2014 n- and I communicate with it ...

HOwever when I am running Instrument Manager - i add the scope, but then I have this message.
Please help
BOBAH
 
Posts: 3
Joined: Sun Dec 18, 2011 3:55 am

Re: TDS3014B connecting with and RS232 to USB adapter

Postby cball on Mon Dec 19, 2011 4:28 pm

Which instrument manager are you referring to? Did you put both NI VISA and Tek VISA on your machine?
Tektronix Application Engineer
Anything I post here is user to user advice. For official technical support:
http://www.tek.com/webform/technical-support-request
cball
 
Posts: 575
Joined: Fri Dec 17, 2010 11:17 pm
Location: Beaverton, OR

Re: TDS3014B connecting with and RS232 to USB adapter

Postby BOBAH on Tue Dec 20, 2011 12:56 am

I ma referring to Tektronix Instrument Manager Software, version 2.6, built Feb 20, 2002

I think I installed both NI VISA and Tek VISA. But how can I double check it?

P.S. How can I attach the screenshots to this post?
BOBAH
 
Posts: 3
Joined: Sun Dec 18, 2011 3:55 am

Re: TDS3014B connecting with and RS232 to USB adapter

Postby cball on Tue Dec 20, 2011 3:24 pm

This sounds like the instrument manager for Wavestar. This can work with NI VISA (it is the configuration I use). I would recommend making sure you have the latest version of NI VISA (5.1.1) and the latest version of Wavestar (3.0). Do no install (or uninstall any existing) anything reference to Tek VISA or Open Choice, to ensure that you do not have two visas installed. Wavestar comes packaged with a version of TekVISA, so if you do not opt to no install it, it will put Tek VISA on your machine.

Note - You cannot currently attach pictures in the forums, but you can give links to other sites where the picture is hosted.
Tektronix Application Engineer
Anything I post here is user to user advice. For official technical support:
http://www.tek.com/webform/technical-support-request
cball
 
Posts: 575
Joined: Fri Dec 17, 2010 11:17 pm
Location: Beaverton, OR

Re: TDS3014B connecting with and RS232 to USB adapter

Postby Tonderon on Wed Jan 25, 2012 6:42 am

Thanks for letting me know all this good stuff. I will take it into consideration to fix the same problem with adaptares. I will check this out!!!
Tonderon
 
Posts: 1
Joined: Tue Jan 24, 2012 9:45 am

Re: TDS3014B connecting with and RS232 to USB adapter

Postby BOBAH on Fri Mar 23, 2012 2:12 am

dear chall,

Actually now I have to VISA's. Tek VISA , and NI VISA. When I go to NI MEeasurement and Automation Explorer -> Devices and Interfaces - I have VISA error. ANd also my serial port does not work.
What shall I do? How can I uninstall Tek VISA and fix the problem.
BOBAH
 
Posts: 3
Joined: Sun Dec 18, 2011 3:55 am

Re: TDS3014B connecting with and RS232 to USB adapter

Postby cball on Fri Mar 23, 2012 3:42 pm

Uninstalling the TekVISA is the path I would recommend. From you will want to use the add/remove programs from the Control Panel in Windows (called Programs and Features in Win 7) and uninstall OpenChoice TekVISA and TekVISA USB-TMC. I would then go to NI's website and download an reinstall the latest version of NI-VISA (I belive 5.1.1 is the most recent version as of today). This should correct the conflict.
Tektronix Application Engineer
Anything I post here is user to user advice. For official technical support:
http://www.tek.com/webform/technical-support-request
cball
 
Posts: 575
Joined: Fri Dec 17, 2010 11:17 pm
Location: Beaverton, OR