TDS 2024 GPIB measured value different to that on screen.

Moderators: Hiker, notnomis, Buck, Super Mod

TDS 2024 GPIB measured value different to that on screen.

Postby ChristinaJ on Fri Jun 29, 2012 5:18 pm

I have a TDS 2024.

When I try to read channel 2 via GPIB I get a ceiling value (E+ 37) and not the value on the scope screen.

Channels 1 & 3 can be read with no problem.

Does anyone know why this may be?

Many thanks for any help you can give,
Christina
ChristinaJ
 
Posts: 3
Joined: Fri Jun 29, 2012 5:08 pm

Re: TDS 2024 GPIB measured value different to that on screen.

Postby cball on Fri Jun 29, 2012 5:44 pm

Which measurement command are you using to query the result?
Tektronix Application Engineer
Anything I post here is user to user advice. For official technical support:
http://www.tek.com/webform/technical-support-request
cball
 
Posts: 575
Joined: Fri Dec 17, 2010 11:17 pm
Location: Beaverton, OR

Re: TDS 2024 GPIB measured value different to that on screen.

Postby ChristinaJ on Fri Jun 29, 2012 5:53 pm

cball wrote:Which measurement command are you using to query the result?


MEASURE:MEAS2:VALUE?

I'm reading each channel seperately.
ChristinaJ
 
Posts: 3
Joined: Fri Jun 29, 2012 5:08 pm

Re: TDS 2024 GPIB measured value different to that on screen.

Postby cball on Fri Jun 29, 2012 7:21 pm

The MEAS:MEAS<X> commands refer to the on screen measurements. The <x> here indicates which measurement you are working with however, not which channel to make the measurement on. So, if I wanted to setup the on screen measurements for CH2 amplitude, as the only on screen measurement, then I would do the following:

MEASU:MEAS1:TYPE AMP
MEASU:MEAS1:SOU CH2
MEASU:MEAS1:STATE ON

-trigger the scope here and make sure you have a good acquisition

MEASU:MEAS1:VALUE?


One trap that you can fall into is that the MEASU:MEAS commands are updated periodically by the scope. If you trigger the scope, then immediately query the MEASU:MEAS1:VALUE?, it might still have and return the measured value from the last acquisition. If you use the MEASU:IMM commands instead, it will calculate the value on the fly for you, and return the results so that you know it is from the current acquisition. There is only 1 MEAS;IMM slot though, so if you need to do several measurements, you will have to change the MEAS:IMM:TYPE between each query.
Tektronix Application Engineer
Anything I post here is user to user advice. For official technical support:
http://www.tek.com/webform/technical-support-request
cball
 
Posts: 575
Joined: Fri Dec 17, 2010 11:17 pm
Location: Beaverton, OR

Re: TDS 2024 GPIB measured value different to that on screen.

Postby ChristinaJ on Mon Jul 02, 2012 5:06 pm

Many thanks for your help.

I managed to fix the problem by shifting my external trigger pulse in time. I think it was a synchronisation problem and I was somehow trying to read the value too soon after the trigger.
ChristinaJ
 
Posts: 3
Joined: Fri Jun 29, 2012 5:08 pm